Sonography Principles and Instrumentation Questions and Answers
Question 29
Which index is related to the likelihood of cavitation?
Options:
A.
Thermal
B.
Temporal
C.
Mechanical
D.
Acoustical output
Answer:
C
Explanation:
The Mechanical Index (MI) is related to the likelihood of cavitation, which is the formation of gas bubbles in a liquid due to the low-pressure regions of the ultrasound wave. MI is a parameter that predicts the potential for mechanical bioeffects, including cavitation. A higher MI indicates a greater likelihood of cavitation occurring. It is calculated based on the peak negative pressure and the frequency of the ultrasound wave.
ARDMS Sonography Principles and Instrumentation guidelines
Kremkau, F. W. (2015). Diagnostic Ultrasound: Principles and Instruments.
Question 30
Which artifact results from decreased attenuation?
Options:
A.
Comet tail
B.
Enhancement
C.
Ringdown
D.
Reverberation
Answer:
B
Explanation:
Enhancement is an artifact that results from decreased attenuation. When an ultrasound wave travels through a medium with lower attenuation compared to surrounding tissues, it loses less energy. Consequently, the structures located deeper than the low-attenuation medium appear brighter on the ultrasound image. This artifact is commonly observed behind fluid-filled structures, such as cysts or the urinary bladder, where the sound waves encounter minimal resistance and thus less attenuation.References:
ARDMS Sonography Principles and Instrumentation guidelines
"Diagnostic Ultrasound: Principles and Instruments" by Frederick W. Kremkau
Question 31
Which function is the purpose of the matching layer?
Options:
A.
Convert acoustic to electrical energy
B.
Allow sequential firing of elements
C.
Reduce ringing of the crystals
D.
Increase sound transmission
Answer:
D
Explanation:
Comprehensive and Detailed Explanation From Exact Extract:
The matching layer is placed between the transducer’s piezoelectric element and the patient’s skin. Its primary function is to improve the transmission of sound energy into the body by reducing the acoustic impedance mismatch between the crystal and tissue, minimizing reflection at the transducer surface.
According to sonography instrumentation reference:
“The matching layer improves transmission of ultrasound energy by reducing impedance mismatch between the transducer and soft tissue, thus increasing the efficiency of sound energy transfer.”
Therefore, the correct answer is D: Increase sound transmission.
—
Question 32
Which adjustment can maintain the same frame rate when the depth is increased?
Options:
A.
Increase number of focal zones
B.
Decrease image width
C.
Increase frequency
D.
Decrease persistence
Answer:
B
Explanation:
When the depth of imaging is increased, the time it takes for the ultrasound pulses to travel to and from the deeper structures also increases, which can reduce the frame rate. To maintain the same frame rate, one effective adjustment is to decrease the image width. Narrowing the image width reduces the number of scan lines required to create each frame, allowing the system to maintain a higher frame rate despite the increased depth.
ARDMS Sonography Principles and Instrumentation guidelines
Kremkau, F. W. (2015). Diagnostic Ultrasound: Principles and Instruments.